The medical gas outlet
mamdouh khairy
16 فبراير 2014
The medical gas outlet shall be an Amico Alert-1 series.Outlets shall be manufactured with a 6 - 1/2” [165 mm] length
type “K” 1/2” [12.7 mm] outside diameter (3/8” nominal) size
copper inlet pipe stub, which is silver brazed to the outlet
body. The body shall be of 1-5/16” [33 mm] diameter, one
piece brass construction. For positive pressure gas services,
the outlet shall be equipped with a primary and secondary
check valve and the secondary check valve shall be rated at
a maximum of 200 psi [1,379 kPa] in the event the primary
check valve is removed for maintenance. Outlet bodies shall
be gas specifi c by indexing each gas service to a gas specifi c
dual pin indexing arrangement on the respective identifi cation
module.
A large color coded front plate shall be used for ease of gas
identifi cation and aesthetic appeal.
A one piece chromed fascia plate shall frame the outlet. With
the back rough-in mounted, the outlet shall adjust from 3/8”
[10 mm] to 1” [25 mm] variation in wall thickness.
The outlets shall be of modular design and include a gas
specifi c 16 ga. [1.6 mm] steel mounting plate designed to
permit on-site ganging of multiple outlets, in any order, on
5” [127 mm] spacing.
The latch-valve assembly shall be Ohmeda DiamondTM (Ohio
Medical) Quick Disconnect compatible and accept only corresponding
Ohmeda type gas specifi c adapters.
All M.R.I. outlets are manufactured from nonferrous materials.
All outlets shall be cleaned and degreased for medical gas
service, factory assembled and tested.
The outlets are UL listed and CSA approved.
Amico products comply with NFPA-99.